Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Agilent TapeStation Software Displays a Different ScreenTape Expiration Date
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • -Tapestation
      • Discrepancy between Agilent RIN and RINᵉ values
      • How to Determine the DV200 of FFPE RNA Samples on the Agilent TapeStation
      • Caught Exception in Agilent 2200 TapeStation Controller Software: Configuration System Failed to Initialize
      • Agilent TapeStation Software Displays a Different ScreenTape Expiration Date
      • Peak Fronting in NGS Library Smears on Agilent TapeStation
      • Agilent 4150 TapeStation Lid does not Close
      • Intensity Loss for 200 nt Ladder Peak in the Agilent TapeStation RNA Assays
      • How to Export Agilent TapeStation Logs
    • +Parallel Electrophoresis

Agilent TapeStation Software Displays a Different ScreenTape Expiration Date

Created by Geoff Pettitt Geoff Pettitt 2 months ago | Last modified by Geoff Pettitt Geoff Pettitt 1 month ago

This Information Applies To: 4200, 4150 and 2200 TapeStation Systems


Issue

The expiration date displayed in the Agilent TapeStation controller software is different to the one printed on the ScreenTape device.


Resolution 

The expiration date printed on the ScreenTape device applies until the device is removed from its packaging. Once a ScreenTape device is removed from its protective plastic sleeve, it dries out quicker. As the ScreenTape device dries out, bubbles can form in the gel, which can hinder sample migration. Therefore, a new expiration date is required. The expiration date shown in the software is the correct one, as it supersedes what is printed on the ScreenTape device. 

 The 4200 and 4150 TapeStation controller software calculates a new expiration date 14 days from when a ScreenTape device is placed in the ScreenTape nest and scanned for the first time. The ScreenTape device does not have to be used for a new expiration date to be calculated, the TapeStation instrument must simply read the barcode.

 The 2200 TapeStation system is slightly different as it calculates a new expiration date 14 days from when the ScreenTape device is used in a run for the first time. If the ScreenTape device is only placed in the ScreenTape nest and scanned, a new expiration date won't be calculated as the ScreenTape device must be used in an experiment.

 Agilent does not guarantee the performance of expired ScreenTape devices or reagents and by default the controller software will not allow the use of expired ScreenTape devices. However, it is possible to enable the use of expired ScreenTape devices within the controller software. In the 4200/4150 TapeStation controller software, click the cog in the top-right corner and select Settings. In the settings menu, click the button to toggle the use of expired ScreenTape devices between Yes and No (Figure 1). When using the 2200 TapeStation system, if an expired ScreenTape device is inserted, a pop-up will appear with the message 'ScreenTape has expired. Running expired ScreenTape and reagents is not supported.'. If the user wishes to use the expired ScreenTape device, they can click OK and they can start the run.


  Figure 1. Allowing the use of expired ScreenTape devices in the 4200 and 4150 TapeStation controller software

 Caution 
Caution: The ScreenTape device ID and information about its previous use is stored on the laptop. If a new laptop is connected to the TapeStation instrument, or the existing laptop is re-imaged, the device history will be lost, and any previously used ScreenTapes will be recognized as new.  

 Caution 
Caution: If the date and time settings on the PC are incorrect, the ScreenTape device expiration date will be recorded incorrectly. If the date and time on the PC are changed between a ScreenTape devices first use subsequent uses, the device could be flagged as expired when it isn't.

 Tip 
Learn how to effectively operate your TapeStation System:
Agilent 4200 TapeStation System Manual
Agilent 4150 TapeStation System Manual
Agilent 2200 TapeStation User Manual

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8648

  • software
  • 4150-tapestation
  • how-to-article
  • 2200-tapestation
  • automated-electrophoresis
  • TapeStation
  • 4200-tapestation
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help