      • DV200 Analysis of FFPE RNA Samples with the 2100 Bioanalyzer System
      • Analysis of Yeast RNA on the Agilent Bioanalyzer
      • RIN Not Calculated or NA for Total RNA Samples in the Agilent Bioanalyzer 2100 Expert Software
      • Toggle X-Axis Description From Time (Seconds) to Size for 2100 Bioanalyzer System Data
      • Agilent 2100 Bioanalyzer System Chip not Detected
      • Agilent Bioanalyzer Assay Chip Guide
      • How to Install the Agilent 2100 Expert Software and Connect a PC to the Agilent Bioanalyzer
      • Maintenance of the Agilent 2100 Bioanalyzer System Chip Priming Station
      • Can the Agilent 2100 Bioanalyzer be Used to Determine Size of RNA?
      • Connecting Two Agilent Bioanalyzers to One PC
      • Agilent 2100 Bioanalyzer System Run-time Error 50003 upon 2100 Expert Software Launch
      • Region Information not Included in Agilent 2100 Bioanalyzer System PDF Report
      • How to Run the Hardware Diagnostics on the Agilent Bioanalyzer
      • Incorrect Text Placement and Image Size in Agilent Bioanalyzer 2100 Expert Software PDF Reports
    • +Tapestation
    • +Parallel Electrophoresis

DV200 Analysis of FFPE RNA Samples with the 2100 Bioanalyzer System

Created by Geoff Pettitt Geoff Pettitt 4 months ago | Last modified by Geoff Pettitt Geoff Pettitt 2 months ago

This Information Applies To: 2100 Bioanalyzer, 2100 Expert software


Issue

The DV200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE) samples established by Illumina Inc. in 2016. It represents the percentage of RNA fragments larger than 200 nucleotides relative to all RNA fragments. The DV200 of FFPE RNA samples can be calculated within the Agilent 2100 Expert software.


Steps to follow

Software version B.02.10 or higher 

If you are using 2100 Expert software version B.02.10 or higher, the DV200 RNA Nano and Pico assays will be available to run from the Assay Selection drop-down list (Figure 1). The software version can be displayed in the 2100 Expert software by clicking Help>About 2100 Expert.... If you are using an older version, you can run the DV200 assays by upgrading to the latest software version. The latest software is available to download free of charge here and this article explains how to install it. The DV200 RNA Nano and Pico assays will then be available in the software. An example DV200 region in the 2100 Expert software is shown in Figure 2.


Figure 1. Selecting the DV200 assays from the Assay Selection drop down menu

 
Figure 2. An example DV200 region taken from the Demo DV200 RNA Nano file

Software version lower than B.02.10

If you are using an earlier software version and are unable to upgrade to the latest software version, download the DV200 assay files:

  • DV200 RNA Nano
  • DV200 RNA Pico

Extract and store the DV200 RNA Nano.xsy or DV200 RNA Pico.xsy file in the 2100 Expert Software RNA assay folder: C:\Program Files (x86)\Agilent\2100 bioanalyzer\2100 expert\assays\RNA. The RNA Nano or DV200 RNA Pico assays will then be available in the Assay Selection drop-down list (Figure 2). 

Applying the DV200 calculation to a previously run data file

If you have 2100 Expert software version B.02.10 or higher installed or have saved the DV200 assay files, the DV200 calculation can be applied to a previously run data file:

  1. Open the data file that you want to apply the DV200 to with the 2100 Expert software.
  2. Navigate to the Assay Properties tab.
  3. Click Import Setpoints. A dialog box will open.
  4. Select the DV200 RNA Nano or DV200 RNA Pico .xsy file, depending on the data file open (Figure 3). The DV200 assay parameters will be applied to the data file.  


Figure 3. Importing the DV200 Setpoints into an existing file

 Tip 
Learn how to effectively maintain and operate your Bioanalyzer:
Simplified DV200 Evaluation with the Agilent 2100 Bioanalyzer System
Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide
Agilent RNA 6000 Nano Kit Quick Start Guide
Agilent RNA 6000 Pico Kit Quick Start Guide

 For Research Use Only. Not for use in diagnostic procedures.

PR7000-8566

  • Operation
  • Bioanalyzer Instruments
  • Troubleshooting
  • 2100 Bioanalyzer Laptop
  • Bioanalyzer
  • Bioanalyzer Software
  • 2100 Bioanalyzer
  • 2100 Expert Software
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help