Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Analysis of Single Stranded DNA on the Fragment Analyzer
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

Analysis of Single Stranded DNA on the Fragment Analyzer

Created by Shweta.S Shweta.S 7 months ago | Last modified by Shweta.S Shweta.S 1 month ago

This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer systems, RNA kit (15 nt) (p/n DNF-471-0500/1000), HS RNA kit (15 nt) (p/n DNF-472-0500/1000), Small RNA kit (p/n DNF-470-0275)


Issue

Which kit should be used to analyze single stranded DNA (ssDNA) on the Agilent Fragment Analyzer?


Resolution 

Agilent does not offer any kit dedicated or validated for ssDNA analysis. However, given that RNA assays are validated for RNA, which is a single stranded nucleic acid, RNA kits can be considered. However, please note that the migration profile might be different for DNA versus the validated RNA, so data for ssDNA may not be accurate. Therefore, Agilent cannot guarantee data obtained for ssDNA using RNA kits. If users do decide to use RNA kits for the analysis of ssDNA, then it would be recommended to include ssDNA fragments of known sizes to use as internal standards.

Check this application note, which discusses the analysis of ssDNA molecules for use in CRISPR/Cas HDR.

 Tip 
Learn how to effectively maintain and operate your Fragment Analyzer:
5200, 5300, and 5400 Fragment Analyzer System Manual
ProSize Data Analysis Software User Manual
Agilent DNF-471 RNA kit (15 nt) Quick Guide for Fragment Analyzer Systems
Agilent DNF-472 HS RNA (15 nt) Kit Quick Guide for Fragment Analyzer Systems
Agilent DNF-470 Small RNA kit Quick Guide for Fragment Analyzer Systems user manuals on Agilent.com 

For Research Use only. Not for diagnostic purposes.

PR7000-8636

  • 5200
  • Parallel Electrophoresis
  • 5400
  • Fragment Analyzer Systems
  • Fragment Analyzer
  • 5300
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help