Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Run Aborted and Target Z Position Error on Parallel Capillary Instruments
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

Run Aborted and Target Z Position Error on Parallel Capillary Instruments

Created by Shweta.S Shweta.S 7 months ago | Last modified by Shweta.S Shweta.S 1 month ago

 

This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Agilent Femto Pulse, Agilent ZAG DNA Analyzer


Issue: The instrument encounters a run error, Run Aborted, Position fault: Did not reach target Z position (Figure 1). This error can occur if the drawer gets opened when the stage is still moving. The error occurs soon after starting the run. 

   

Figure 1. Run Error:CtrlTaskStage Error 


Resolution: 

System Reboot: 

Perform an entire system reboot as described in the Agilent Knowledge Portal article Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments and as mentioned in the following steps: 

  1. Turn off the instrument.
  2. Turn off the computer.
  3. Unplug the two USB cables labeled CCD and HUB from the back of the computer.
  4. Wait for approximately 10 minutes before reconnecting the USB cables.
  5. Reconnect USB cables to different, nonadjacent ports and reconnect power cables to the instrument and PC. We recommend keeping at least one port empty (unused) between the two USB cables.
  6. Turn on the computer and wait for it to complete startup.
  7. Turn on the instrument.
  8. Launch the controller software.

If the issue persists, please perform a stage reset as described in the following procedure.

Reset Stage:

Please ensure the instrument is connected and that the stage is not holding any plate. This can be checked by opening each drawer and making sure all the white plate carriers are in the drawers. Then perform a stage reset from the controller software:

  1. From the instrument controller software, open the Utilities drop-down menu (Figure 2).
  2. Select the Hardware I/O option and locate the stage section of the newly created window (Figure 3).
  3. Click Reset stage and ensure that the holding number directly above the reset button goes to zero (Figure 3). Before you reset the stage, please ensure that the stage is not holding a plate, then click Yes to proceed (Figure 4).

     

Figure 2: Select Hardware I/O from the Utilities menu of the controller software. The Fragment Analyzer controller software is shown as an example in this figure.  

    

Figure 3: Reset stage to holding number value to zero.

      

Figure 4: Confirm that you wish to proceed with the stage reset after checking that the stage is not holding a tray.

 If the error persists: 

  1. Take a screenshot of the error message.
  2. Bypass the error(s) and collect the event and error reports from the instrument controller software one by one. To collect the reports, open the instrument controller software (Login as an Administrator), click Admin, select and collect the Event Report, then select the Error Report (Figure 5). 
  3. Please send the screenshots, event report and error report to your local Agilent support representative.

       

Figure 5: Steps to collect the event and error reports. The Fragment Analyzer controller software is shown as an example in this figure. 

 

 Caution 
Caution: Before you reset the stage, ensure that it is not holding a plate. If the stage is holding a plate, the instrument might be damaged and the stage will not get reset. 

 

 Tip 

Learn how to effectively maintain your Agilent parallel capillary electrophoresis systems:
5200, 5300 and 5400 Fragment Analyzer User Manual 
Femto Pulse User Manual 
ZAG DNA Analyzer User Manual 
Oligo Pro II User Manual
Prosize Data Analysis Software User Manual

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8669

  • Troubleshooting
  • femto-pulse
  • ZAG
  • fragment-analyzer
  • parallel-capillary-electrophoresis
  • troubleshooting-article
  • how-to-article
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help