Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument

Created by Shweta.S Shweta.S 7 months ago | Last modified by Shweta.S Shweta.S 30 days ago

This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer and Oligo Pro II instruments, and their respective controller software.


Issue
Failed to build pressure error comes up shortly after starting a separation.


Resolution

Failure to build pressure error occurs when the system is trying to build pressure when liquid is being pumped during conditioning or flushing of the instrument. Possible causes and suggestions to address the issue are: 

1. Empty reagent bottles - As first step, please ensure that there is adequate volume in each reagent bottle and that the respective tubing is in the liquid.

2. Leak - Please check whether there is any visible leak at the tubing (any visible liquid inside the instrument). Figure 1 shows various red tube fittings and other components of the side compartment of the fragment analyzer instrument. The points are going to be similar for other instruments as well. Check for leaks in these areas including but not limited to what is visible in the picture. 

  • If there is any obvious and easily accessible leak, for example around the array bundle, reinstall it into the reservoir (Figure 1, (1)). If there is liquid at the F port (Figure 1, (4)) try to fix by tightening the red tube fittings. 
  • If leak is from the tube that connects to the reservoir, then take a picture (Figure 1) and contact your local Agilent Support from Contact us page. This connection is not usually something users can tighten themselves so an onsite visit by Agilent field engineer will likely be needed.  
  • If there is no leak and issue is not leak related, please proceed to step 3.


Figure 1. Fragment Analyzer side door compartment.


Various leak points are: (1) Top of the array bundle (2) Vent valve (3) F-port connections, located by red ferrules (4) Conditioning line (5) syringe. 

3. Reboot - Perform a full system reboot following, see the steps in the article: Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments. Full power cycle is recommended to address any possible component disconnects. If that resolves the issue, it is recommended to monitor for some time. If not, please proceed to step 4.

4. Pump related - To determine whether issue is pump-related, the recommendation is to execute a Conditioning Solution flush so that pump moves few times. Click Add to queue, select Full Conditioning or any method using Conditioning Solution and edit it to only contain the Conditioning step. Click OK twice to add it to the queue. (Figure 2).  During the flush step, please monitor syringe and liquid movement. 

    

Figure 2. Adding Capillary Array Conditioning to queue. 

  • Syringe movement indicates that the pump is working.
  • If pump does not move, then the issue is likely instrument hardware failure. An onsite visit by an Agilent engineer will be needed. Therefore, please collect and share instrument Event and Error reports (Figure 3) from the instrument controller software (Admin > Error or Event Report > choose at least dates at which the error occurred) with your local Agilent support from Contact us page. 


Figure 3.  Steps to collect Event and Error reports from controller software

 Tip 
Learn how to effectively use your parallel capillary electrophoresis instruments:
5200, 5300 and 5400 Fragment Analyzer User Manual 
Femto Pulse User Manual 
ZAG DNA Analyzer User Manual 
Oligo Pro II User Manual

For Research Use only. Not for use in diagnostic procedures.

PR 7000-8759

  • Troubleshooting
  • femto-pulse
  • ZAG
  • fragment-analyzer
  • oligo-pro-II
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help