Automated Electrophoresis Articles How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments

Created by Geoff Pettitt Geoff Pettitt 9 months ago | Last modified by Shweta.S Shweta.S 1 month ago

This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer


Issue:

Dirt or dust on the capillary array window can cause baseline shifts or elevation of the baseline. Figure 1 shows an example electropherogram with baseline shifts. It is sometimes possible to see the dirt or dust in the View Capillary Positions screen in the Agilent ProSize data analysis software. This screen can be opened by clicking Analysis>View Capillary Positions in ProSize. An example of dust on a capillary array is shown in Figure 2.

Figure 1. A portion of a Fragment Analyzer electropherogram with Baseline shifts indicated with blue arrows.   

Figure 2. Fragment Analyzer capillary positions from the View Capillary Positions window in ProSize. The capillaries are depicted as distinct blue peaks. Dust will show as additional signal, see right-most peak.

If the baseline shift or elevation has occurred in just one run, the instrument ventilation between runs may be sufficient to remove the particles. However, if the baseline shifts or baseline elevation persists for multiple runs cleaning the capillary array window can be a useful troubleshooting step. 


Steps to Follow:

 Caution 
Caution: Be gentle to avoid damage to the capillaries or capillary array.
  1. Open the side door and hood of the instrument.
  2. Remove the light guide from the array window as shown in Figure 3.
  3. Use a small nylon paintbrush or Kimwipe to gently clean the dust off the window while the window is dry. Brush across the window from left to right or right to left, not up and down. The window is highly fragile! Figure 4 demonstrates the capillary array window cleaning using a paintbrush on an Agilent Fragment Analyzer. The dust is typically on the capillaries due to static cling and can be removed easily with this step. If more intensive cleaning is needed, proceed to steps 4–9. Otherwise, proceed to step 10.
  4. Remove the bundle end of the capillary array using the capillary array bundle removal tool. Place bundle in provided protective cover.
  5. Remove the capillary array window from the capillary array window holder. Do not touch the array window.
  6. Place a paper tower behind the capillary array window. An example of a Fragment Analyzer or ZAG DNA Analyzer capillary array window is shown in Figure 5 and a Femto Pulse capillary array window is shown in Figure 6.
  7. Using a spray bottle filled with 70% isopropanol or ethanol solution, gently spray the capillary array window.
  8. Use a small nylon paintbrush to gently brush the capillaries in one direction while they are still wet. Alternatively, a lint-free cloth can be used to blot the array window dry.
     Caution 
    Caution: It is important to let the capillaries air dry before re-attaching the light guide. The light guide can evaporate the alcohol solution resulting in alcohol condensing on the glass filter behind the array window.
  9. Reinstall the capillary array window, bundle, and light guide.
  10. Perform a separation on the instrument. This separation should have sample, marker, or blank solution in every capillary position so that the file can be used to check the capillary alignment. 
  11. Check the alignment of the capillaries in the Agilent ProSize data analysis software. Click Analysis>View Capillary Positions. In the View Capillary Positions window, if the alignment is correct, the red dots will be at the top of each blue peak. If the alignment does not look correct, it is possible to drag and drop the red dots onto the top of the blue peaks. Once you are happy with the capillary alignment, click Accept to confirm the current capillary positions and return to the main ProSize menu. This will correct the alignment for this data file only.
  12. If you had to adjust the capillary alignment for this separation, then it will be necessary to correct the alignment using the instrument controller software. Capillary alignment in the controller software will avoid the alignment being incorrect in future data files. Please check Chapter 4 of the Fragment Analyzer System Manual, Femto Pulse User Manual or ZAG DNA Analyzer System Manual for instructions. It is recommended and easiest to follow the Capillary Alignment from a File method. The capillary alignment should be done with the new data file, as only this latest run will have the current capillary positions.

  Figure 3. The two white screws that secure the light guide to the window being removed on a Fragment Analyzer


Figure 4. Cleaning a Fragment Analyzer capillary array window

 

Figure 5. Fragment Analyzer or ZAG DNA Analyzer capillary array window with paper towel behind

Figure 6. Femto Pulse capillary array window with paper towel behind

 Tip 
Learn how to effectively operate your Agilent parallel capillary electrophoresis system:
5200, 5300, and 5400 Fragment Analyzer System Manual
Femto Pulse User Manual
ZAG DNA Analyzer System Manual 
ProSize Data Analysis Software User Manual

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8238

  • ZAG DNA Analyzer
  • Troubleshooting
  • Femto Pulse
  • Automated Electrophoresis
  • Capillary Arrays
  • Parallel Electrophoresis
  • Fragment Analyzer
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help