Automated Electrophoresis Articles How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems

Created by Geoff Pettitt Geoff Pettitt over 1 year ago | Last modified by Shweta.S Shweta.S 8 months ago

This Information Applies To: 5200, 5300, 5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, and ZAG


Issue:

Adding mineral oil to Agilent parallel capillary electrophoresis DNA marker plates and samples plates prevents evaporation and enables re-use of the plates. 

 Caution 
Caution: Agilent has not validated adding mineral oil to RNA plates. It is used at the customers own risk. 

Steps to follow:

 Tip 
Tip: Agilent recommends using mineral oil part number FS-SMO15 to overlay sample or marker plates when appropriate.
  1. Add one drop of mineral oil into each well of the prepared marker or sample plate directly from the mineral oil drop bottle. One drop is equivalent to a volume of 20 µL. See the Quick Guide supplied with kit you are running for information on how soon the mineral oil covered plate should be used.
Tip

Learn how to effectively operate your Agilent Parallel Capillary Electrophoresis system: 
5200, 5300, and 5400 Fragment Analyzer System Manual
Femto Pulse User Manual
Oligo Pro II User Manual 
ZAG DNA Analyzer System Manual 

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8119

  • Fragment Analysis Kits and Reagents
  • ZAG DNA Analyzer
  • Operation
  • Femto Pulse
  • Best practices
  • Oligo Pro II Reagents
  • Parallel Electrophoresis
  • Femto Pulse Analysis Kits and Reagents
  • Fragment Analyzer
  • ZAG DNA Analysis Kits and Reagents
  • Oligo Pro II
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help