This Information Applies to: Agilent 2100 Bioanalyzer system, RNA 6000 Nano kit, RNA 6000 Pico kit, and Small RNA kit
Issue
The Bioanalyzer RNA Ladder and/or samples are degraded.
Background
Two examples of degraded RNA Ladders are shown in Figure 1.
Figure 1. Electropherogram of a partially (left) and fully (right) degraded Bioanalyzer RNA Ladder.
Steps to Follow
RNA degradation can occur either before or during the chip preparation.
Degradation before the Bioanalyzer runs
The samples and ladder may have been degraded before loading onto the Bioanalyzer chip. To confirm if the ladder or samples were already degraded, consider checking their quality by another method such as with the Agilent TapeStation or Fragment Analyzer systems.
Degradation during chip preparation
If the samples or ladder were not degraded before loading on the Bioanalyzer chip, then they may have been contaminated with RNase during the chip preparation or chip run. The source of RNase contamination must be removed. Perform the following steps before running a new chip:
General precautions to avoid RNA degradation
Kit | Before each run | After each run | Monthly or after liquid spill: pin set cleaning |
RNA Nano | Electrode cleaner:
|
Electrode cleaner:
|
RNase decontamination with brush:
|
RNA Pico and Small RNA | Electrode cleaner:
|
Electrode cleaner:
|
RNase decontamination with brush:
|
Table 1. Maintenance of the electrode cartridge
Learn how to effectively maintain and operate your Agilent 2100 Bioanalyzer system: 2100 Expert Software User's Guide on Agilent.com Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.