This Information Applies to: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer systems
Issue
One or more electropherograms have baseline noise.
Background
An example of baseline noise is shown in Figure 1.
Figure 1. Unexpected peaks and baseline noise in a ZAG DNA Analyzer electropherogram caused by bubbles in the reservoir.
Steps to Follow
There are several causes of parallel capillary electrophoresis baseline noise:
After considering the preceding troubleshooting steps, run another separation to check if the baseline has improved. If these steps have not helped, contact Agilent support. To find the correct contact, go to the Contact Us page and choose your country.
Learn how to effectively maintain your Agilent Fragment Analyzer, Femto Pulse, or ZAG DNA Analyzer system: 5200, 5300 and 5400 Fragment Analyzer system manual on Agilent.com Femto Pulse user manual on Agilent.com ZAG DNA Analyzer system manual on Agilent.com |