This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, Oligo Pro II, ZAG DNA Analyzer and ProteoAnalyzer systems
Issue
The Agilent parallel capillary electrophoresis assay quick guides contain the guidance to allow reagents to equilibrate at room temperature before use. What is considered room temperature?
Steps to Follow
For the parallel capillary electrophoresis assays, allow your reagents to equilibrate to 19 to 23 oC for 30 minutes before use.
Learn how to effectively maintain and operate your Agilent parallel capillary electrophoresis instrument:
5200, 5300, and 5400 Fragment Analyzer System Manual on Agilent.com
Femto Pulse User Manual on Agilent.com
Oligo Pro II User Manual on Agilent.com
ZAG DNA Analyzer System Manual on Agilent.com
ProteoAnalyzer System Manual on Agilent.com ProSize Data Analysis User Manual on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.