Atomic Absorption Articles How to Resolve SpectrAA Software Issues due to Adobe Flash Failure
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • +Atomic Absorption Operation
    • +Atomic Absorption Maintenance
    • -Atomic Absorption Troubleshooting
      • Agilent SPS 4 Autosampler Initialization Sequence and Troubleshooting
      • Agilent SPS 3 Autosampler Initialization Sequence and Troubleshooting
      • Agilent SPS 3, SPS 4 Autosampler Offline Message
      • How to Resolve SpectrAA Software Issues due to Adobe Flash Failure
    • +Atomic Absorption Best Practices

How to Resolve SpectrAA Software Issues due to Adobe Flash Failure

Created by Tina Harville Tina Harville over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: Agilent Atomic Absorption systems with SpectrAA Software 5.x


Issue:

Following Adobe Flash going out of support as of 31 December 2020, Adobe has disabled the Adobe Flash Player as of 12 January 2021. SpectrAA uses Adobe Flash for the start screen.

This procedure outlines the steps to resolve SpectrAA locking up on initialization due to Adobe Flash Failure.


Steps to follow:

  1. The SpectrAA configuration can be changed to prevent the execution of Adobe Flash content using a text editor. However, as this configuration file is normally protected, the text editor must be run using administrator privileges. To do this procedure, first type notepad into the search box on the Windows taskbar.

  2. Click the down arrow to show more options for opening the notepad text editor and click Run as administrator


  3. From within the notepad text editor, click File > Open


  4. Navigate within the Open dialog to the location:

    C:\Program Files(x86)\Agilent\SP100\Run\HELP (if running 64-bit version of Windows) or C:\ Program Files\Agilent\SP100\Run\HELP (if running 32-bit version of Windows)

  5. Click the drop-down to change from showing Text Documents (*.txt) to All Files (*.*).

  6. Open the file IndexPage.ini. Change the line SWF=Yes in this document to read SWF=No.

  7. Click File and then Save. Close the Notepad text editor. When next started, the SpectrAA application will revert to using the non-Adobe Flash application window.

  • SpectrAA
  • Troubleshooting
  • AAS Software
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help