Atomic Absorption Articles Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • +Atomic Absorption Operation
    • -Atomic Absorption Maintenance
      • Atomic Absorption: Resetting Hollow Cathode Lamp Usage Counters
      • Replacing the PSD 120 Syringe and Plunger
      • General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems
      • Agilent AA Atomizer Disassembly and Re-assembly
      • How to clean a blocked nebulizer of an Agilent Flame Atomic Absorption System
      • Flame AA Burner Maintenance
      • How to Change a Deuterium (D2) Lamp
      • Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
      • Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components
      • Troubleshooting and Maintenance of AA Systems: Part 2B- Flame Optimization
      • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization
      • VGA Maintenance and Cleaning
      • SIPS 10/20 Maintenance
    • +Atomic Absorption Troubleshooting
    • +Atomic Absorption Best Practices

Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization

Created by Whitney Blackmon Whitney Blackmon over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: Agilent Graphite Furnace Atomic Absorption and Zeeman Instruments (240/280(Z) AA)


Issue:

How to optimize the Instrument to allow for better results, troubleshooting, after maintenance or lamp replacement.


Steps to follow:

 Warning 
Warning: The furnace runs at temperatures up to 3000 degree Celsius. Allow furnace to cool completely before touching.

Optimize one or more lamps: 

Lamp optimization is required after cleaning or installing a new lamp or used lamp. Lamp optimization can identify when a lamp has reached the end of its working life. 

  1. Remove the furnace workhead from the light path before optimizing the lamps: 
    1. Tilt the Zeeman workhead forward, or
    2. Release the GTA thumbscrew (Figure 3) and remove the workhead. 

  2. Load a method and select Optimize button > OK to open the Furnace Optimization window. While watching the green Gain % bar (Figure 1), adjust the two knobs at the base of the HCL lamp (Figure 2).


    Figure 1. Lamp Optimization

    Figure 2. Hollow Cathode Lamp Adjustment

  3. The lamp is optimized when the intensity (green bar) is in scale and at the highest point. If the bar is green to the top, select Rescale and continue optimizing. Once the lamp is fully optimized, record the Gain % value. 

     Tip 
    Tips: Each time the lamp is optimized the Gain % value will be close to the previous value and the original value. A significant difference indicates that the lamp is near or at the end of its life. See Troubleshooting and Maintenance of Atomic Absorption Systems: Part 1 - Hollow Cathode Lamps - YouTube
  4. Move the workhead back into the light path: 

    1. Tilt the Zeeman workhead backwards, or

    2. Replace the GTA workhead and lock down the thumbscrew (Figure 3). 


      Figure 3. Furnace Workhead Adjustment. 1 GTA thumbscrew location 2 Horizontal adjustment knob 3 Vertical adjustment knob

  5. While watching the green Gain % bar (Figure 1), optimize the horizontal and vertical position of the workhead using the appropriate knob (Figure 3). The workhead is position is optimized when the intensity (green bar) is in scale and at the highest point. The vertical and horizontal adjustments can be performed in any order:

  6. Turn on the water chiller/exchanger and argon gas supply.

  7. Replace the Agilent Programmable Sample Dispenser / Autosampler (PSD) on the front of instrument with the two hooks in sample compartment. Tighten the alignment block on the right side of the PSD-120. Align the PSD. See Alignment of the Sample Probe on the Agilent PSD 120. 

  8. Monitor the injection of a blank solution and calibration standard. The absorbance valus for the blank (less than 0.01 Abs) and standard should return to normal operating conditions. 

 Tip 
Learn how to effectively operate your Agilent Atomic Absorption System:
Troubleshooting and Maintenance of Atomic Absorption Systems: Part 3 - Graphite Furnace AAS YouTube
Agilent 240/280 Series AA User's Guide manual on Agilent.com




  • SpectrAA
  • aa software
  • 240z
  • atomic absortion
  • zeeman
  • GFAA
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help