Atomic Absorption Articles Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • +Atomic Absorption Operation
    • -Atomic Absorption Maintenance
      • Atomic Absorption: Resetting Hollow Cathode Lamp Usage Counters
      • Replacing the PSD 120 Syringe and Plunger
      • General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems
      • Agilent AA Atomizer Disassembly and Re-assembly
      • How to clean a blocked nebulizer of an Agilent Flame Atomic Absorption System
      • Flame AA Burner Maintenance
      • How to Change a Deuterium (D2) Lamp
      • Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
      • Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components
      • Troubleshooting and Maintenance of AA Systems: Part 2B- Flame Optimization
      • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization
      • VGA Maintenance and Cleaning
      • SIPS 10/20 Maintenance
    • +Atomic Absorption Troubleshooting
    • +Atomic Absorption Best Practices

Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps

Created by Tina Harville Tina Harville over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: All Agilent AA systems: 55B, 240FS, 280FS, 240Z, 280Z


Issue:

Optimizing the position of the lamp maximizes amount of light from the lamp passing through the optics to the detector, improving the signal-to-noise and the instrument detection limit. 


Steps to follow:

  1. Check the recommended current and wavelength for the lamp, and check the selections in the software. This information is recorded on the label near the base of the lamp.

  2. On the Analysis page, select Optimize > OK. 

     Tip 
    Tip: Record the gain value the first time you optimize the lamp. Each time you optimize the lamp, the value should be close to the noted value. Track the gain value over time to monitor the performance of the lamp. A significant change in the gain can indicate that the lamp is reaching the end of its operational life.
  3. Slowly rotate the lamp adjustment screws at the base of the lamp, while watching the lamp optimization bar and numerical display. Generally, only small adjustments will need to be made. Record the optimized gain value.



  4. If the display goes off-scale, click Rescale and continue optimizing until the maximum value is obtained on the display.

  5. If you have a flame instrument, and you enabled the background correction, you will see a second optimization bar for the deuterium background correction lamp. Simply rotate the deuterium lamp adjustment screws, located at the bottom left of the lamp compartment. Watch the optimization bar and numerical display and adjust the screws to achieve the maximum value. Only minor adjustments will normally be required. Full re-alignment of the deuterium lamp is usually required only when replacing the lamp.



  • AAS Instruments
  • Troubleshooting
  • 240FS AA
  • 240Z AA
  • 280FS AA
  • 55B AA
  • Maintenance
  • 280Z AA
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help