Atomic Absorption Articles Agilent SPS 4 Autosampler Initialization Sequence and Troubleshooting
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • +Atomic Absorption Operation
    • +Atomic Absorption Maintenance
    • -Atomic Absorption Troubleshooting
      • Agilent SPS 4 Autosampler Initialization Sequence and Troubleshooting
      • Agilent SPS 3 Autosampler Initialization Sequence and Troubleshooting
      • Agilent SPS 3, SPS 4 Autosampler Offline Message
      • How to Resolve SpectrAA Software Issues due to Adobe Flash Failure
    • +Atomic Absorption Best Practices

Agilent SPS 4 Autosampler Initialization Sequence and Troubleshooting

Created by Paul Gunston Paul Gunston 11 months ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler


Issue: When the Agilent SPS 4 autosampler powers up, it performs an initialization routine, which moves the probe arm to predefined positions on all its axes (X, Y, and Theta (rotational). See Figure 1. If the probe arm fails to travel the required distance along any of the axes, LEDs will light to indicate an error. The autosampler cannot be used in an error state.

Figure 1. Axis directions.   1 error LEDs,  2 X-axis,  3 Theta (T)-axis, 4 Z-axis

Discussion:  

The initialization sequence is as follows (in this order):

  • probe raises to the extremes of the Z-axis
  • probe moves to front-middle position
  • probe travels to the extremes of the X-axis
  • probe travels to the extremes of the theta-axis
  • probe is moved around its motion envelope (eight positions)
  • probe positions itself above the fixed wash reservoir position (rear-middle)


Steps to Follow: If the autosampler fails to complete the initialization.

 Warning 
Warning: The probe arm moves quickly. Keep the probe path clear during operation.
  1. Turn off the power and check each axis path manually to determine if there is any physical or mechanical obstruction.
  2. If no obstruction can be identified, turn the unit back on and allow it to attempt the reset procedure again.
  3. If still failing, the Error LEDs on the autosampler connection panel on the front of the instrument will display a code. There are four LED lights. 

The LED pattern, where 1 is on and 0 is off, corresponds to an error code and description (Table 1). Note the error code and contact your Agilent Service Representative.

LED 1

LED 2

LED 3

LED 4

Error code

Error description

0

0

0

1

1

DSID Error

0

0

1

0

2

Program Memory Checksum Error

0

0

1

1

3

Configuration Memory Checksum Error

0

1

0

0

4

RAM Test Error

0

1

0

1

5

Axis Initialization Error

0

1

1

0

6

X-Axis Collision Error

0

1

1

1

7

Theta-Axis Collision Error

1

0

0

0

8

Z-Axis Collision Error

1

0

0

1

9

EEPROM Error

1

0

1

0

10

Mat Checksum Error

1

0

1

1

11

X-Axis Movement Error

1

1

0

0

12

Theta-Axis Movement Error

1

1

0

1

13

Z-Axis Movement Error

1

1

1

0

14

Reserved

1

1

1

1

15

CPU Exception Error

Table 1: Error codes and description

 Tip 
Learn how to effectively troubleshoot your Agilent Spectroscopy System: Guidelines for Troubleshooting and Maintenance of AA Systems

 

  • sps4
  • samplers
  • atomic absortion
  • AA Samplers
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help