Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Atomic Absorption Articles AA software - SpectrAA Flags
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • -Atomic Absorption Operation
      • How to Set up the PSD Probe Arm
      • Alignment of the Sample Probe on the Agilent PSD 120
      • Flame Optimization for Agilent Atomic Absorption Systems
      • AA Calibration Curves Types
      • AA software - SpectrAA Flags
      • VGA Setup
      • Shutting Down the Agilent Vapor Generation Accessory (VGA)
    • +Atomic Absorption Maintenance
    • +Atomic Absorption Troubleshooting
    • +Atomic Absorption Best Practices

AA software - SpectrAA Flags

Created by Paul Gunston Paul Gunston 11 months ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: Agilent SpectrAA Software


Issue: 

Explanation of concentration results flags from SpectrAA Software


Under certain conditions, a concentration result may appear with a concentration result 'Flag'. The following table lists each possible concentration result flag and its description. The Priority column indicates the priority number of each flag, Priority 1 = highest priority. If a sample concentration results in more than one error being flagged, the flag with the highest priority is displayed.

The flag may simply be noted for reference purposes or may indicate that corrective action is needed, or one or more standards or samples may need to be rerun.

Flag

Priority

Flag Description

A

6

For Zeeman, the absorbance has exceeded the maximum Zeeman absorbance value set in the Method. Absorbances above the maximum value give erroneous results due to Zeeman calibration rollover.

B

11

The absolute concentration of the solution is less than RDL (Required Detection Limit) and greater than or equal to IDL (Instrument Detection Limit).

D

3

Autodilution error. Volume is less than 1 µL for furnace; factor is greater than 50 for SIPS.

G

4

Low mains voltage.

I

9

Atomic signal absorbance is less than minimum absorbance (user specified limit).

L

13

LCS (Laboratory Control Sample) is out of limits.

M

10

%RSD (percent Relative Standard Deviation) is greater than limit.

N

13

Matrix spike recovery is outside limits.

Q

13

QC (Quality Control) Standard recovery is outside limits.

T

5

Total absorbance is over-range.

U

12

The absolute concentration of the solution is less than IDL.

P

13

Preparation blank absolute concentration is greater than user specified limit. See Note #2.

W

13

QC Spike percent recovery is outside upper and lower limits.

Y

13

The RDL concentration is outside the user specified limits. See Note #2.

Z

13

QC Blank absolute concentration is greater than or equal to the user specified limit.

*

13

Duplicate Relative Percent Difference exceeded.

+

13

Correlation coefficient less than the specified limit.

a

7

Foreground absorbance is over-range. See Flag "a" Note.

e

2

This data item has been edited.

b

8

Background absorbance is over-range (>2 ABS).

m

1

The solution has been measured in read mode.

p

13

PROMT (Precision Optimized Measurement Time) error. This error indicates that the instrument detected a zero or negative transmittance value.

r

13

The solution has been remeasured.

?

13

There are also 10 user definable tests available. The flag for each test is also user-defined. (only available for SpectrAA Pro or CFR (full QCP) software)

Tip
Learn how to effectively maintain and operate your Agilent Spectroscopy System:
Agilent 240/280 Series AA Users Guide




  • SpectrAA
  • software
  • aa software
  • atomic absortion
  • Data Analysis
  • AA
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help