      • How to Set up the PSD Probe Arm
      • Alignment of the Sample Probe on the Agilent PSD 120
      • Flame Optimization for Agilent Atomic Absorption Systems
      • AA Calibration Curves Types
      • AA software - SpectrAA Flags
      • VGA Setup
      • Shutting Down the Agilent Vapor Generation Accessory (VGA)
    • +Atomic Absorption Maintenance
    • +Atomic Absorption Troubleshooting
    • +Atomic Absorption Best Practices

How to Set up the PSD Probe Arm

Created by Whitney Blackmon Whitney Blackmon over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent AA (AA240, AA280) with Graphite Furnace Accessory (GTA)


Issue:

The Programmable Sample Dispenser (PSD) requires the sampling probe to be accurately aligned before the PSD can be used.


The PSD enables the automation of the analysis of samples with the furnace.

When in use, the PSD sits in the front sample compartment of the AA, with the carousel containing the required solutions for analysis.

When the PSD is set up, it is locked into position to maintain alignment. The position of the PSD can be adjusted both in/out or sideways to align the probe arm accurately to the furnace workhead.

The position of the capillary tubing will need to be set up to ensure that:

  • The capillary is correctly located in the vials when picking up solution.
  • The capillary enters and leaves the GTA without touching the sides of the injection hole in the workhead.

Steps to follow:

1. To set up the probe arm, thread the capillary tubing into the capillary holder following the instructions below. When properly inserted it should appear as shown in Figure 1 and Figure 2.

Figure 1 - Illustration of probe arm showing capillary tubing

To thread the tubing:

  1. Undo and remove the brown knurled probe guide screw.
  2. Thread a suitable length of capillary through the threading hole.
  3. Thread the tubing through the O-rings in the probe guide collar.
  4. Feed the tubing until it protrudes from the base of the probe guide collar
  5. Tighten the knurled probe guide screw
     

Figure 2- Photo of capillary tubing in place.

2. The capillary holder is then adjusted so that the capillary is central to, and not touching, the bottom of the vial. The probe arm should be central in the injection hole in the workhead when the probe arm moves to this position, see Figure 3 and Figure 4. For adjustment, use the adjustment knobs located under the carousel, see Figure 5).

 

Figure 3 - Video showing correct adjustment of the capillary holder.

Figure 4 - Probe positioning in a sample vial

Figure 5 - PSD showing adjustment and locking knobs 

 Tip
Tips: Avoid misalignment by always performing adjustments with a straight capillary

The PSD is now aligned.

  • atomic absortion
  • zeeman
  • AA
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help