Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Atomic Absorption Articles Agilent AA Atomizer Disassembly and Re-assembly
  • Atomic Absorption Articles
  • Atomic Absorption Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Atomic Absorption
    • +Atomic Absorption Operation
    • -Atomic Absorption Maintenance
      • Atomic Absorption: Resetting Hollow Cathode Lamp Usage Counters
      • Replacing the PSD 120 Syringe and Plunger
      • General Maintenance for the Agilent Flame and Graphite Furnace Atomic Absorption Systems
      • Agilent AA Atomizer Disassembly and Re-assembly
      • How to clean a blocked nebulizer of an Agilent Flame Atomic Absorption System
      • Flame AA Burner Maintenance
      • How to Change a Deuterium (D2) Lamp
      • Troubleshooting and Maintenance of AA Systems: Part 1 - Hollow Cathode Lamps
      • Troubleshooting and Maintenance of AA Systems: Part 2A- Cleaning Flame AA Components
      • Troubleshooting and Maintenance of AA Systems: Part 2B- Flame Optimization
      • Troubleshooting and Maintenance of AA Systems: Part 3- Graphite Furnace Optimization
      • VGA Maintenance and Cleaning
      • SIPS 10/20 Maintenance
    • +Atomic Absorption Troubleshooting
    • +Atomic Absorption Best Practices

Agilent AA Atomizer Disassembly and Re-assembly

Created by Jo Boden Jo Boden over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA55, AA240, AA280) 


Issue:

How to disassemble and re-assemble the atomizer (burner, spray chamber, nebulizer)


Steps to follow:

 Warning 
Warning:
  • The burner operates at high temperature. Allow sufficient cooling time before handling.
  • There may be traces of acid on surfaces of the atomizer. Wear appropriate personal protective equipment.

1. Ensure that the flame is extinguished and the burner has cooled. Remove the burner and spray chamber (see Figure 1).

Figure 1 - Removing the burner and spray chamber

2. Prepare the following three tools: 1. Venturi Extraction tool (black), 2. Nebulizer Tool (silver) and 3. Flat-blade screwdriver (yellow handle). These tools are included with the AA when purchased. See Figure 2 below. 

Figure 2 - Venturi Extraction Tool (black, top left), Nebulizer Tool (silver, top right) and Flat-blade screwdriver (bottom) 

3. Use a flat-blade screwdriver to loosen the impact bead (glass bead) setscrew. Remove the glass bead. See Figure 3.  

Figure 3 - Loosen the glass bead setscrew and remove the glass bead

4. Loosen the nebulizer uptake adjustment (black) knob and use the nebulizer tool (silver) to loosen the capillary retaining nut (rotate anticlockwise). See Figure 4. 

Figure 4 - Loosening the nebulizer uptake adjustment knob (black) and the capillary retaining nut (rotate anticlockwise)

5. Remove the atomizer needle retaining ring, capillary, and spring.

6. Use the venturi extraction tool to push the remaining parts from the front. Cover the exit end of the chamber before pushing out to avoid losing parts). See Figure 5. 

Figure 5 - Using the venturi extraction tool to push remaining parts out of the spray chamber

7. Clean the removed nebulizer components. Components can be cleaned with water with a dilute detergent in an ultrasonic bath for 10 minutes. Worn or damaged O-rings (items 5,7,9) should be replaced. All required O-rings are available as a kit under Agilent part number 9910093400. For a parts breakdown, see Figure 6. 

Figure 6 - 1. Thimble 2. Locking ring 3. Clamp bush guide 4. Capillary assembly 5. O-ring 6. Spring 7. O-ring 8. Capillary guide 9. O-ring 10. Venturi 

8. After cleaning, place the nozzle and capillary tube into the nebulizer, then use the venturi extraction tool to position it to the correct placement (all the way in). See Figure 7. 

Figure 7 - Installing the nozzle and capillary tube into the nebulizer

9. Place the spring, capillary, and capillary retaining ring in sequence. Use the nebulizer tool to lock the atomizing needle retaining ring. See Figure 8. 

Figure 8 - Installing the spring, capillary, and capillary retaining ring, then locking the atomizing needle retaining ring

10. Finally, lock the flow adjustment knob.

11. Before installing the impact bead, move the impact bead mount to the closest adjustment point and install the glass bead.

12. Make sure that there is a gap (~1–2 mm) between the glass bead and the venturi (nozzle) before locking the setscrew.

13. Re-assemble the drain tank to the spray chamber and refit to the burner adjuster. Refit the burner.

The maintenance of the atomizer is now complete.

 Warning 
Warning: 
  •  Incorrect nebulizer assembly and fitting in an atomic absorption spectrometer can create an explosion and fire hazards, which can cause serious personal injury and damage to property
  • Check that the nebulizer is correctly assembled and is properly adjusted as described in the Operation Manual. Gas connections should be tested for leaks, and any leaks fixed before lighting the flame.

 Tip 
Learn how to effectively maintain AA Spectroscopy System: Guidelines for Troubleshooting and Maintenance of AA Systems; Spectroscopy Supplies



  • SpectrAA
  • Maintenance
  • how-to-article
  • flame aa
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help