How to perform a cold start for 1200 series LC modules

Document created by lcdk Employee on Aug 15, 2017Last modified by lcdk Employee on Sep 3, 2017
Version 3Show Document
  • View in full screen mode

In some situations, it is necessary to perform a so-called cold start (forced coldstart). Reasons can be commands which have not been correctly processed in the electronics, or sensors that have become obsolete during operation. With this cold start, the module will be reset to the factory settings.

 

Step-by-step guide

 

Processing on Infinity I systems

 

Switch off the module

 

On the back of the module, make a note of the position of the "dip switches" (see the red marking in the drawing; example pump)

 

Switch the "dip switches" to the following position (position 1 is above, position 0 is below). For the cold start switch "dip switches" 1, 2 and 8 up - the others (3-7) down.

 

 

Switch the module on again and wait until the initialization is finished (status LED off).

 

Switch the module off again and reset the original position of the "dip switches". Then switch the module on again. Cold start is completed and the module can be controlled again.

 

 

Processing on Infinity II systems

 

Switch off the module

 

On the back of the module, make a note of the position of the "dip switches" (see the red marking in the drawing; example VWD)

 

Switch the "dip switches" to the following position (position 1 is above, position 0 is below). For the cold start switch "dip switches" 1 and 6 up - the others (2-5) down.

 

 

Switch the module on again and wait until the initialization is finished (status LED off)

 

Switch the module off again and reset the original position of the "dip switches". Then switch the module on again. Cold start is completed and the module can be controlled again.

 

 

Additional Read

LabAdvisor Manual

2 people found this helpful

Attachments

    Outcomes