Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV

Document created by agilent Employee on Jun 15, 2016
Version 1Show Document
  • View in full screen mode

When using an Isocratic pump or a Binary pump without a Solvent Selection Valve (SSV), you can utilize more than 1 degasser chamber to achieve higher degassing performance and help the degasser work optimally.

 

This is accomplished with the C-Tube Kit (part G1379-68706) to link two degassing chambers together.  For example, chambers A+B can be connected and C+D can be connected.

 

This will add some volume when changing solvents, but will not lead to any delay volume in gradient runs since the gradient starts at the pump.

 

Some methods benefit more than others, such as running Methanol+Water in lower wavelength areas where extra degassing helps (see Document "Micro Degasser Based Baseline Wander on UV-Detector Signals for very Sensitive Measurements in the Low UV-Wavelength Range".

 

This document was generated from the following discussion: Use of Multiple Degasser Chambers for Isocratic and Binary pumps without SSV