Autostart macro failed error 999 in OpenLAB CDS ChemStation

Version 3

    Autostart Macro failed, error# 999 occurred (OpenLAB CDS Chemstation Rev. C.0x.0x)

     

    This is usually resolved by resetting the instrument configuration files in the instrument folder (\Chem32\x\, where x is the number of the instrument in question).

     

    Append .old to these files with all sessions (online and offline) of ChemStation turned off:

    config.reg

    ChemStationPerInstrumentSettings.config

    instrument.config

    RapidControl.InstrumentConfig.xml

     

    After modification, the files will read:

    config.reg.OLD

    ChemStationPerInstrumentSettings.config.OLD

    instrument.config.OLD

    RapidControl.InstrumentConfig.xml.OLD

     

    Start ChemStation again and the problem should be resolved.