No MS Data Found - G1701FA Chemstation Drug Analysis.

Document created by agilent Employee on May 23, 2016
Version 1Show Document
  • View in full screen mode

As seen in EA versions of Chemstation, when a data file is corrupt and data analysis has a problem starting gives the error message

'No MS Data found'

Data Analysis will not open.

Changing the file name to a default file such as Evaldemo.d will correct the problem. Or you can just rename files.val to files.org and it will rebuild it.

 

Locate the data path for evaldemo.d for example, 

d:\masshunter\gcms\data\evaldemo.d

 

Locate the files.val file

c:\masshunter\gcms\1

 

[last files]

data1=D:\data\20140707\001SP0001.D

shortdata1=001SP0001.D

data2=D:\data\5977-1_002.D

shortdata2=5977-1_002.D

data3=D:\data\5977-1_001.D

shortdata3=5977-1_001.D

 

point the path to evaldemo.d

[last files]

data1=D:\data\evaldemo.d

shortdata1=evaldemo.d                    

data2=D:\data\evaldemo.d

shortdata2=evaldemo.d

data3=D:\data\evaldemo.d

shortdata=evaldemo.d

 

save and restart data analysis.

Or point to the directory path where evaldemo.d exists.

 

 