Normal Phase seals and wear-in procedure

Document created by agilent Employee on May 19, 2016Last modified by agilent Employee on May 19, 2016
Version 2Show Document
  • View in full screen mode

There is a wear-in procedure in the 1200/1260 pump manuals.  This wear in procedure is for reverse phase seals, not normal phase seals.  The manual states in the description of the Seal wear in procedure that normal phase seals would be damaged by the procedure.  The pressure being over 300 bar damages the normal phase seals.   IPA is fine with the normal phase seals.

 

 