How to ... scale time axis to show the time range between 2 named reference peaks - OpenLAB CDS Intelligent Reporting

Document created by agilent Employee on Mar 11, 2016Last modified by agilent Employee on Mar 28, 2016
Version 2Show Document
  • View in full screen mode

Q: A user wants to scale the chromatogram's time axis from 10% left of reference peak 1 to 10% of reference peak 2. The name of the reference peaks is given.

 

A:

  • Generate a group including a chromatogram plot   (add a dummy filed so you can generate a group around the plot)
  • Repeat the group on a suitable parameter (in order to allow filtering)
  • Filter the group on the names of the 2 reference compounds
  • Scale the time axis using min/max values of peak begin/peak end.

 

See the attached screenshots for further details.

2012-1026_scaling.PNG

By the way – you can repeat a group on the expression “=1” …

Just to get filtering.

 

(originally posted in OpenLAB Forum: https://zohodiscussions.com/openlabcdsforum#Topic/68030000000046014

2012 October 26

by betrich)