How to ... refer to compound specific constants stored in a larger table

Document created by agilent Employee on Mar 28, 2016
Version 1Show Document
  • View in full screen mode

With OpenLAB CDS 2.0 it will be possible to refer to an external constant file in order to use externally stored parameters in calculations or reports.

 

With OpenLAB CDS A versions you could follow a 2 step approach to embed such a list of compound specific parameters into your report template:

 

1. Define an internal report parameter to store the constants. The name/value pairs need to be entered as a single long string, where the name and values are separated by e.g. a semicolon.   (Benzene;98.0;Toluene;97.5;.....)

 

2.  Define a group repeat based on compound names and extract the constant from the string and store it in a variable with "compound name" as a unique key value

 

Syntax to find  the compound specific start position of recovery value in the parameter string:

=InStr(Parameter1,Compound_Name)+len(Compound_Name)+1

 

Syntax to extract the compound specific recovery value from start position to the next 4 characters.

=(Mid(Parameter1,(InStr(Parameter1,Compound_Name)+len(Compound_Name)+1,4)

 

In order to use the attached example template you require a method with identified or calibrated peaks named Benzene and Toluene (mind the spelling!). This template can be used with OpenLAB CDS A.01.05 or higher.

 

(originally posted in OpenLAB Forum: https://zohodiscussions.com/openlabcdsforum#Topic/68030000000186001

2015 June 3 by Bernhard Etrich)

 