How to.... do bracketing calibration calculation within the report template (OpenLAB CDS Intelligent Reporting)

Version 2

    Setting up reports on bracketed calibrations could be a challenge as calibration runs belong to multiple result groups and the association to 2 result groups cannot be achieved by e.g category aggregators.

     

    The following describes a simple but effective workaround:

    • Introduce a sample related custom field called “bracket ID”. Use a text type custom field.
    • Do not use the ChemStation’s bracketing feature but set up the sequence in a ‘rolled’ out manner, like
      STD, Sample 1, Sample2, Sample 3,STD, Sample 4, Sample 5, Sample 6, STD, Sample 7, Sample 8, Sample 9,STD,…
    • Fill the bracket ID custom field for each sequence line. Standards that belong to 2 brackets should get 2 labels, e.g. comma separated:

    Sample      BracketID

    STD                       01

    Sample 1              01

    Sample 2              01

    Sample 3              01

    STD                       01,02

    Sample 4              02

    Sample 5              02

    Sample 6              02

    STD                       02,03

    Sample 7              03

    Sample 8              03

    Sample 9              03

    STD                      03

     

    • In the report template you could use a CONTAINS filter condition and apply it to the custom field.
    • Generate a result table for bracket 1, including all calculations. When everything is done, copy the result table and change the filter condition to bracket ID 2 and so on.

     

    Unfortunately, there is no FOR/NEXT functionality to loop the result table based on the bracket ID. However, the setup is not too complicated in case of 5 or 6 brackets per sequence.

     

     

    (originally posted in OpenLAB Forum: https://zohodiscussions.com/openlabcdsforum#Topic/68030000000021019

    2012 July 10

    by Bernhard Etrich