How to.... do bracketing calibration calculation within the report template (OpenLAB CDS Intelligent Reporting)

Document created by agilent Employee on Mar 11, 2016Last modified by agilent Employee on Mar 28, 2016
Version 2Show Document
  • View in full screen mode

Setting up reports on bracketed calibrations could be a challenge as calibration runs belong to multiple result groups and the association to 2 result groups cannot be achieved by e.g category aggregators.

 

The following describes a simple but effective workaround:

  • Introduce a sample related custom field called “bracket ID”. Use a text type custom field.
  • Do not use the ChemStation’s bracketing feature but set up the sequence in a ‘rolled’ out manner, like
    STD, Sample 1, Sample2, Sample 3,STD, Sample 4, Sample 5, Sample 6, STD, Sample 7, Sample 8, Sample 9,STD,…
  • Fill the bracket ID custom field for each sequence line. Standards that belong to 2 brackets should get 2 labels, e.g. comma separated:

Sample      BracketID

STD                       01

Sample 1              01

Sample 2              01

Sample 3              01

STD                       01,02

Sample 4              02

Sample 5              02

Sample 6              02

STD                       02,03

Sample 7              03

Sample 8              03

Sample 9              03

STD                      03

 

  • In the report template you could use a CONTAINS filter condition and apply it to the custom field.
  • Generate a result table for bracket 1, including all calculations. When everything is done, copy the result table and change the filter condition to bracket ID 2 and so on.

 

Unfortunately, there is no FOR/NEXT functionality to loop the result table based on the bracket ID. However, the setup is not too complicated in case of 5 or 6 brackets per sequence.

 

 

(originally posted in OpenLAB Forum: https://zohodiscussions.com/openlabcdsforum#Topic/68030000000021019