Replacing the Syringe Carriage Assembly in the Agilent 7693A Automatic Liquid Sampler

Document created by carlos_vargas Employee on Oct 14, 2020Last modified by carlos_vargas Employee on Oct 23, 2020
Version 6Show Document
  • View in full screen mode

Agilent Support Insights

 

This Information Applies To: Agilent 7693A Automatic Liquid Sampler

 

Issue:

This document explains how to replace the syringe carrier assembly in case you want to change from standard to high volume syringes, or if the Agilent support representative determines the assembly has to be exchanged.

 

 

Use the standard syringe carriage assembly for syringe sizes up to 100 μL, and the Enhanced Sample Handling Syringe Carriage for syringe sizes from 250 μL to 500 μL, see Figure 1.

 

Standard Syringe Carrier Assembly - Agilent

Standard Syringe Carriage Assembly

Enhanced Syringe Carrier Assembly - Agilent

Enhanced Sample handling Syringe Carriage Assembly

Figure 1. Syringe Carriage Assemblies for the 7693A Autoinjector

 

Steps to follow:

To Remove the Syringe Carriage Assembly:

 

 1. Remove all vials and bottles from the turret, then disconnect the injector cable from the GC. You may need to lift the injector tower in order to disconnect the cable.

 2. If desired, remove the injector from the mounting post and mount the injector tower on a parking post.

 3. Open the injector door and remove the syringe see Removing and Installing a Syringe on the Agilent 7693A Autosampler 

 4. Remove the ALS Turret, see Changing the Turret of the 7693A Automatic Liquid Sampler 

 

5. Slide the syringe carriage assembly down until the assembly cable is accessible          below the tower casing, and detach the cable from the assembly.

6. Slide the syringe carriage assembly completely up until the flange and latch are            accessible.

7. Completely loosen the T-10 Torx screw above the syringe latch.

8. Completely loosen the T-10 Torx screw at the top of the syringe carriage                        assembly.

9. Carefully remove the syringe carriage assembly from the injector carriage.

Remove ALS Syringe Carrier-Agilent

Figure 2. Removing the syringe carriage assembly

 

To Install the Syringe Carriage Assembly:

 

1. Slowly position the replacement syringe carriage assembly on the injector            carriage. The syringe carriage assembly will hang in place when positioned        correctly.

2. Completely tighten the T-10 Torx screw above the syringe latch.

3. Completely tighten the T- 10 Torx screw at the top of the syringe carriage            assembly.

4. Slide the syringe carriage assembly completely down until the cable input jack        on the assembly is accessible below the injector casing.

5. Connect the assembly cable to the syringe carriage assembly.

6. Slide the syringe carriage assembly completely up. 

Install Syringe Carriage ALS-Agilent

Figure 3. Mounting the syringe carriage assembly

 

 7. Reinstall the turret, see Changing the Turret of the 7693A Automatic Liquid Sampler 

 8. Install the syringe and close the injector door, see Removing and Installing a Syringe on the Agilent 7693A Autosampler 

 9. Plug in the injector cable and, if necessary, install the injector tower on the mounting post.

 10. Align the injector, see How to align the Agilent 7693A GC injector 

 11. If you have a sample tray, calibrate the ALS system, see How to Calibrate the Agilent 7693A Sample Tray 

Learn more on how to operate your Agilent 7693A Automatic Liquid Sampler:

SI-7693-1100s - Agilent 7693A ALS Basic and Advanced Operation and Maintenance e-Learning course on Agilent University 

 

 

Keywords: carousel, ALS, GC, GC Samplers, 7693A

 

 