How to Calibrate the Agilent 7693A Sample Tray

Document created by carlos_vargas Employee on Oct 12, 2020Last modified by carlos_vargas Employee on Oct 14, 2020
Version 6Show Document
  • View in full screen mode

Agilent Support Insights

 

This Information Applies To: Agilent 7693A Automatic Liquid Sampler

 

Issue:

The Automatic Liquid Sampler (ALS) system calibration process aligns the sample tray with the injector turret position so that vial transfers occur without incident.

Calibration should be performed if a calibration does not exist, and also as a routine maintenance procedure.

Calibrating the Automatic Liquid Sampler system is recommended if any ALS components are moved.

 

Steps to follow

  1.  Place the calibration vial part number G4514-40588 (Figure 1) in the tray position 1 (bottom right on Figure 2).
    Calibration Vial ALS - AgilentTray position 1 - Agilent
    Figure 1. Calibration Vial G4514-40588Figure 2. Tray position 1
  2. Remove any vials from transfer turret positions L1, L2, and L3 for all installed injectors.
  3. Start the ALS system calibration:
    • Intuvo 9000 GC: In the browser interface, or touch screen, select Settings > Calibration > ALS/Tray > Start Calibration
    • 8890 GC: In the browser interface, or touch screen, select Settings > Calibration > ALS/Tray > Start Calibration
    • 8860 GC: In the browser interface, select Settings > Calibration > ALS/Tray > Start Calibration
    • 7890 Series GC: Use the GC front keypad and select [Options] Calibration > ALS > Start Calibration.
    • 6890 Series GC: Use the GC front keypad and select [Options] Calibration > Sample tray > Start Calibration.
  4. The following calibration process will occur for all installed injectors:
    1. The tray assesses the turret alignment by placing the calibration vial into turret position L1 and returning to sample tray position 1.
    2. The tray tests the vial height and turret position using the alignment tab between positions L1 and L2 on the transfer turret.
    3. The tray verifies the turret alignment by placing the calibration vial into turret position L1 and returning it to sample tray position 1.
  5. When the calibration process is complete, the green Ready status light will come on and the gantry will stop in the Home position.

    Remember to replace any vials that were moved for the calibration process.

 

Learn more on how to operate your Agilent Automatic Liquid Sampler:

SI-7693-1100s - Agilent 7693A ALS Basic and Advanced Operation and Maintenance  e-Learning course at Agilent University 

 

 

Keywords: Tray calibration, vial not aligned, tray not aligned, G4514A, ALS Tray, GC Samplers, GC

 

 